Full 2L QBank
Physiology Circulation 7d61f6cc

Which is true about measurement of BP with sphygmomanometer versus intra-aerial pressure measurements?

A
Less than intravascular pressure
B
More than intravascular pressure
C
Equal to intravascular pressure
D
Depends upon blood flow
High-Yield Explanation
The reason for this is - "dissipation of pressure". When BP is measured by sphygmomanometry, the principle is 'balancing of pressure between the cuff and the brachial aery'. Suppose the pressure in the aery is, say, 100 mm Hg. When the pressure just outside the aery is 100 mm Hg, the aery will be occluded completely, or when the pressure outside the aery becomes just less than 100, the aery will open up and the flow will sta in the aery. However, for pressure just outside the aery to be 100, pressure in the cuff has to be about 106 mm Hg; about 6 mm Hg is dissipated in the intervening tissue (muscle, fat) between the cuff and the aery. Thus, if we are looking for a point at which the aery is just occluded (with aerial pressure being 100), the cuff pressure at which it happens is about 106 mm Hg Hence, by sphygmomanometry, the pressure measured is higher as compared to direct intravascular pressure measurement.

Related Physiology MCQs

Practice 2,00,000+ NEET PG Questions Free

Timed mock tests, mistake queue analytics, audio lectures & zero attempt limits on i❤️Exams.

Start Free Mock Test Now